PB 4-Input, negative values from correctly attached load cell

Post by cheister »

Hi,
I am using the PhidgetBridge 4-Input and a 100Kg Load cell, which worked perfectly until some days ago where it startet to measure negative values, which vanished, after restarting the computer it is attached to. What could be the reason for this. The load cell is not exposed to extreme temperature changes, nor is it under continuous load and also used far below its maximum. Many thanks, C.
User avatar
mparadis
Site Admin
Posts: 959
Joined: Fri Oct 28, 2011 12:17 pm
Contact:

Re: PB 4-Input, negative values from correctly attached load cell

Post by mparadis »

Did the values gradually drift into the negatives or was it a sudden change? Are these the raw values from the voltageRatio event, or is there some kind of math being applied to them in your program?

You should also double check the terminal blocks to make sure there's a solid connection on all four wires, since a loose connection could cause strange values like this and could be "fixed" if the board is moved in just the right way.
cheister
Phidgetsian
Posts: 12
Joined: Wed Jul 21, 2021 6:20 am
Contact:

Re: PB 4-Input, negative values from correctly attached load cell

Post by cheister »

Thanks for your reply. It is the raw value read from the phidgets tray control. The values jumped and were normal again after a restart of the host computer. The connections looked fine, but I will double check them.
User avatar
mparadis
Site Admin
Posts: 959
Joined: Fri Oct 28, 2011 12:17 pm
Contact:

Re: PB 4-Input, negative values from correctly attached load cell

Post by mparadis »

It's hard to know for sure what happened, but hopefully it was just an issue with the connection of one of the wires or something. If it happens again, try to capture any sensor data or logs you can, and that will help us figure out what could be causing it.
cheister
Phidgetsian
Posts: 12
Joined: Wed Jul 21, 2021 6:20 am
Contact:

Re: PB 4-Input, negative values from correctly attached load cell

Post by cheister »

It seems that the cables were the reason. Although they seemed to be fix before, after loosening and retightening them it worked fine again. Should have checked this before posting it... Thanks again. :-)