Error in Wheatstone Bridge User Guide

Comments & issues
Post Reply
dpetrillo
Fresh meat
Posts: 3
Joined: Tue Jan 23, 2018 3:35 pm
Contact:

Error in Wheatstone Bridge User Guide

Post by dpetrillo »

I have the wheatstone bridge phidget. If I connect a bridge with a fairly big output span ~ 20mV/V, I am getting saturation warnings from the phidget control panel for that channel when I look at the output and squeeze my load cell. The problem is, for 64x and 128x they are off by an order of magnitude.

The user guide here: https://www.phidgets.com/docs/DAQ1500_U ... al_Details

says that the range should be ±78.125 mV/V and ±39.063 mV/V for 64 and 128x gain respectively. I am finding it to be 7.8 and 3.9 mV/V. Sure enough, if you apply a gain of 128 to the unamplified range of 500 mV/V you get 3.9 mV/V...

Seems like an error to me.
User avatar
mparadis
Site Admin
Posts: 648
Joined: Fri Oct 28, 2011 12:17 pm
Contact:

Re: Error in Wheatstone Bridge User Guide

Post by mparadis »

You are correct, good catch. I've corrected the user guide.