problems reading resistance (bridge - 1046)

Posted: Wed Apr 05, 2017 10:00 am
by diogenis
I am connecting one of the channels of the bridge in full bridge configuration, using 3x 1k precision resistors. Then I proceed with the calibration using 100ohms and 1K also precision.
Under windows the program finds the equation and carries out the identification of 100&1k resistors fine, but when I try the 10k resistor it reports junk. Why?

Re: problems reading resistance (bridge - 1046)

Posted: Thu Apr 06, 2017 7:29 am
by mparadis
I believe this is because a wheatstone bridge circuit can only correctly measure a resistance up to a maximum of the resistance of it's other resistors. Since you're using 1K resistors, it will start to give incorrect readings after a resistance of 1K. If you're using load cells or strain gauges, their internal resistance is usually 1K or less.